My multimeter (Axiomet AX-105) after years of reliable work started to show incorrect unstable values of low resistance. The reason was - old lubricant over contacts of rotary switch. After cleaning by alcohol a good contact is back and becomes even better than was new - typical 0.1 Ohm value of short circuit now is 0.0 almost all time. But I think that purpose of lubricant is a protection of contacts from abrasion. What is your experience or knowledge? What kind of lubricant do you use as a replacement of aged one? Is it easy accessible? Or do you simply leave it dry without further sad consequences?
|
|
« Last Edit: December 28, 2015, 03:58:43 15:58 by Signal »
|
Logged
|
Give a right name to a right game and play it right
|
|
|
solutions
Hero Member
Offline
Posts: 1826
Thank You
-Given: 656
-Receive: 905
|
|
« Reply #1 on: December 29, 2015, 12:14:42 00:14 » |
|
Any decent meter, including an old analog one with a galvanic movement, can be zeroed out and should be, for leads' and contacts' resistances.
That said, unless someone knows of something better, there's a conductive grease that typically is used to protect lead acid battery terminals from corroding and from acid exposure - I'm guessing that MAY work.
|
|
|
Logged
|
|
|
|
|
Signal
Active Member
Offline
Posts: 200
Thank You
-Given: 113
-Receive: 81
|
|
« Reply #3 on: December 29, 2015, 03:13:25 15:13 » |
|
Any decent meter, including an old analog one with a galvanic movement, can be zeroed out and should be, for leads' and contacts' resistances.
If shorted Ohmmeter indicates form 2 to 120 Ohms depending on light touches to the switch knob and indication changes in time itself that is not the case for trimming/zeroing. That said, unless someone knows of something better, there's a conductive grease that typically is used to protect lead acid battery terminals from corroding and from acid exposure - I'm guessing that MAY work.
Conductive grease spread over dielectric surface of board between exposed contacts looks irrelevant. I am pretty sure - it is not applicable here.
|
|
|
Logged
|
Give a right name to a right game and play it right
|
|
|
dR-mR
Junior Member
Offline
Posts: 57
Thank You
-Given: 59
-Receive: 106
|
|
« Reply #4 on: December 30, 2015, 11:21:47 11:21 » |
|
|
|
|
Logged
|
|
|
|
Signal
Active Member
Offline
Posts: 200
Thank You
-Given: 113
-Receive: 81
|
|
« Reply #5 on: December 30, 2015, 12:13:11 12:13 » |
|
I used that lubricant for my old multimeter without problems ...
I suppose that you use one from reserves on your shelf. How long do you use the multimeter after application of that lubricant?
|
|
|
Logged
|
Give a right name to a right game and play it right
|
|
|
dR-mR
Junior Member
Offline
Posts: 57
Thank You
-Given: 59
-Receive: 106
|
|
« Reply #6 on: December 30, 2015, 12:34:59 12:34 » |
|
3 or 4 months. I gave the multimeter to a friend he works with car electronics, i think he still use the multimeter. The lubricant i got from the same friend
|
|
|
Logged
|
|
|
|
sotos234
V.I.P
Newbie
Offline
Posts: 32
Thank You
-Given: 697
-Receive: 16
|
|
« Reply #7 on: January 02, 2016, 08:43:37 08:43 » |
|
Silicone Dielectric Grease is preferred.
|
|
|
Logged
|
|
|
|
|
|
mylogin
Junior Member
Offline
Posts: 71
Thank You
-Given: 9
-Receive: 16
|
|
« Reply #10 on: January 31, 2016, 05:08:40 05:08 » |
|
Deoxit works great for my multi-meter rotary switch, phone jack, Ethernet jack, RF connector, and power adapter connector contact, especially for old instrument switch contacts. --highly recommended.
|
|
|
Logged
|
|
|
|
|
GroundPlaneLoop
Newbie
Offline
Posts: 19
Thank You
-Given: 3
-Receive: 13
|
|
« Reply #12 on: February 05, 2016, 07:03:43 19:03 » |
|
Isopropanol with a little white mineral oil works also good.
|
|
|
Logged
|
|
|
|
Signal
Active Member
Offline
Posts: 200
Thank You
-Given: 113
-Receive: 81
|
|
« Reply #13 on: January 24, 2017, 12:53:01 12:53 » |
|
I found "KONTAKT CHEMIE Kontakt Gold 2000" as an ideally targeted solution though the cost is higher (14euro/200ml). http://www.crceurope.com/wwwcrc/tds/TKC3%20KoGold2000.pdfGeneral properties and applications:KONTAKT CHEMIE Kontakt Gold 2000 is a long-lasting and temperature-resistant lubricant. The synthetic oil forms extremely thin layers with minimum influence on the contact resistance of electrical contacts. The lubricant has proved to be particularly effective forcontacts with surface coatings of soft metals, e.g. gold, silver, tin. The corrosion-inhibiting effect of noble-metal coatings is improved considerably because they are no longer scratched by mechanical loads and vibrations. KONTAKT GOLD 2000 "Protective lubricant for metal-plated contacts" is referenced in technical data sheet of another product: KONTAKT 61 Contact lubricant & corrosion inhibitor: http://www.crceurope.com/wwwcrc/tds/TKC3%20Kontakt61.pdf1. GENERAL DESCRIPTION Mineral oil based contact cleaner for all types of electrical contacts. Cleans, lubricates, displaces moisture and prevents corrosion. 4. DIRECTIONS <...> To reduce the wear and tear of soft noble metal (coated) contacts, we do recommend the use of our synthetic contact lubricant KONTAKT GOLD 2000. <...>
|
|
« Last Edit: January 24, 2017, 03:24:45 15:24 by Signal »
|
Logged
|
Give a right name to a right game and play it right
|
|
|
Signal
Active Member
Offline
Posts: 200
Thank You
-Given: 113
-Receive: 81
|
|
« Reply #14 on: November 27, 2017, 03:40:46 15:40 » |
|
Just have cleaned and reapply "KONTAKT CHEMIE Kontakt Gold 2000". After 11 months resistance of contacts rapidly becomes much worse than zero - about 30 Ohms changing by tens from rotation to rotation.
|
|
|
Logged
|
Give a right name to a right game and play it right
|
|
|
optikon
Cracking Team
Hero Member
Offline
Posts: 853
Thank You
-Given: 1110
-Receive: 2007
|
|
« Reply #15 on: November 27, 2017, 11:14:12 23:14 » |
|
Just have cleaned and reapply "KONTAKT CHEMIE Kontakt Gold 2000". After 11 months resistance of contacts rapidly becomes much worse than zero - about 30 Ohms changing by tens from rotation to rotation.
OK good to know... perhaps the issue is that the Gold 2000 is just a conductive lubrication but does not address the oxidation that was present which continues to degrade. Were the contacts first treated with something that removes oxidation? Gold 2000 corrosion inhibitor means that it slows / stops additional but didnt address the original existing condition?
|
|
|
Logged
|
I can explain this to you. I can't comprehend it for you.
|
|
|
Signal
Active Member
Offline
Posts: 200
Thank You
-Given: 113
-Receive: 81
|
|
« Reply #16 on: November 27, 2017, 11:46:45 23:46 » |
|
OK good to know... perhaps the issue is that the Gold 2000 is just a conductive lubrication but does not address the oxidation that was present which continues to degrade. Were the contacts first treated with something that removes oxidation?
Multimeter has "gold" contacts on PCB and bronze sliders. There are no signs of corrosion on both. And I used only isopropanol for cleaning that I believe does not clean oxidation. What I noticed before cleaning - grooves in the layer of old lubricant that looked thickened. Maybe I sprayed too much, do not know. Gold 2000 corrosion inhibitor means that it slows / stops additional but didnt address the original existing condition?
It definitely does not address to remove initial corrosion, as I understand description (see above in this thread). The main target of using lubricant was to protect "gold" layer on PCB from abrasion. Ten months of usage before reapplying is not perfect at all but still does not bother me yet. Will see next year
|
|
|
Logged
|
Give a right name to a right game and play it right
|
|
|
papy_bidouille
Newbie
Offline
Posts: 24
Thank You
-Given: 132
-Receive: 22
|
|
« Reply #17 on: November 28, 2017, 08:55:13 08:55 » |
|
I used a lot of faith with succes this type of special grease for conact electric moving made by SIEMENS type WAHLER FET T. I did not have any more or could find some
|
|
|
Logged
|
|
|
|
Gallymimu
Hero Member
Offline
Posts: 704
Thank You
-Given: 152
-Receive: 214
|
|
« Reply #18 on: November 28, 2017, 03:05:40 15:05 » |
|
Just have cleaned and reapply "KONTAKT CHEMIE Kontakt Gold 2000". After 11 months resistance of contacts rapidly becomes much worse than zero - about 30 Ohms changing by tens from rotation to rotation.
It's likely a losing battle. The original plating on the contacts may be gone (I realize you said it was not) and corrosion may set in MUCH more quickly now. Also there could be pitting, or loss of spring force on the contacts as well. I've never had these things work "like new" after cleaning and lube.
|
|
|